Anyone have any protocol or test procedure for validating the AlignRT software post upgrade? I would presume a monthly cal and SRS Cube cal.
Thanks, Jim
Anyone have any protocol or test procedure for validating the AlignRT software post upgrade? I would presume a monthly cal and SRS Cube cal.
Thanks, Jim
Jim,
We use the following approach after every software upgrade with some items being based on the items affected during the upgrade. Tolerances are up to your institution.
Perform a monthly calibration (we use raised plate here but you can use whatever you wish to use for your clinic)
Perform Daily QA
Complete a Isocenter cube calibration
Daily QA is performed and validated against statistical trend of previous Daily QAs
Isocenter cube calibration verification is completed (tolerances 0.01cm or 0.1degree)
Monthly camera only end-to-end test (no linac based imaging done)
Hidden target test using only the camera for phantom positioning followed by WL imaging a 8 gantry/couch positions (tolerance <1.0mm for SRS)
Hidden target test using the cameras for initial setup of the phantom followed by CBCT positioning and a WL imaging a 8 gantry/couch positions (tolerance <0.5mm for SRS)
Includes TG-142tests, TG-147 tests, Linac interface check, and software functionality tests
Static localization
Dynamic localization
Static position vs DICOM coordinates
Camera stability
Obstruction testing
Gating testing
Automated remote positioning
Surface Tools SSD check
ROI propagation
Reference and treatment imaging